This article contains information on maintaining a PurpleAir sensor’s laser counters, which collect particulate matter data. If your sensor is experiencing issues with temperature, humidity, or VOC readings, try the steps in our BME Troubleshooting article. If your sensor isn’t powering, try the steps in our Power Troubleshooting article.
Once your sensor has been in service at the installation location, it becomes exposed to foreign contaminants. We expect the sensor to contact airborne particulate matter. But undesirable contaminants may impede the function of the sensor.
We do not recommend a regular cleaning schedule for PurpleAir sensors, and no preventive maintenance is required. If one of the laser counters is not reporting as expected, is downgraded, reporting high, not working, or spikes erratically, then it is possible that one of the laser counters in your sensor has become congested with debris, an insect, or a spider.
If a single laser counter is consistently reading at or near 0, the issue can be caused by a broken fan in addition to insects or debris.
Cleaning the Laser Counters
DISCLAIMER: If followed improperly, these troubleshooting steps can result in the fans in your device breaking. The only solution to a broken fan is replacement laser counters.
If necessary, you can clean out the laser counters in a device. This cleaning is meant to clear out debris inside the laser counters themselves, and as such you must make sure you vacuum or use compressed air in the correct locations.
Note that increased airflow speed can cause the fan to spin at a much higher rate than recommended, which can damage or break the fans. When cleaning, physically block the fans with a toothpick or something similar.
First, power off the sensor so that the laser counter fans stop spinning. Place a shop vacuum or traditional vacuum hose on the bottom of the sensor for 2-3 minutes. Clear out the laser counter with canned compressed air if further cleaning is necessary. Compressed air does tend to work better.
For PA-II and PA-II-SD sensors, focus on the four collinear holes on each laser counter (see image below). For PA-I edition sensors, focus on the rectangular hole on the side opposite the fan. For the PurpleAir Flex, focus on the rounded rectangular hole on each laser counter. Do not spray into the fans themselves, always use the intake holes
For sensors with two laser counters, the A channel is the laser counter farthest from the USB port. The B channel is the laser counter closest to the USB port.
Optional Steps
The following cleaning is optional, and is not an official part of our troubleshooting steps.
It is possible that debris caught inside a laser counter can persist after cleaning with a vacuum or compressed air. This type of debris is usually something like a spider web. Given that the previous steps were unsuccessful, a more in depth cleaning could solve the issue. The following video outlines how to open up and clean a laser counter:
Power Issues
Although rare, it is possible for power problems to cause erroneous readings. If your sensors readings are still erroneous after cleaning the laser counters, we recommend the following troubleshooting steps:
- Plugging the sensor into a different power outlet
- Using a different power cable
Replacement Laser Counters
Information pertaining to replacing the laser counters in your device can be found in our Sensor (Laser Counter) Replacement community article.
Learn More